Results: Pulse oximeter readings corresponding to 536 blood samples were analyzed. Four of the 6 oximeters tested showed large errors (up to -6.30% mean bias, precision 4.30%, 7.53 ARMS) in estimating saturation when SaO2 was reduced <80%, and half of the oximeters demonstrated large errors when estimating saturations between 80% and 90%. Two of the pulse oximeters tested (Contec CMS50DL and Beijing Choice C20) demonstrated ARMS of <3% at SaO2 between 70% and 100%, thereby meeting International Organization for Standardization (ISO) criteria for accuracy.
Results: Pulse oximeter readings corresponding to 536 blood samples were analyzed. Four of the 6 oximeters tested showed large errors (up to -6.30% mean bias, precision 4.30%, 7.53 ARMS) in estimating saturation when SaO2 was reduced <80%, and half of the oximeters demonstrated large errors when estimating saturations between 80% and 90%. Two of the pulse oximeters tested (Contec CMS50DL and Beijing Choice C20) demonstrated ARMS of <3% at SaO2 between 70% and 100%, thereby meeting International Organization for Standardization (ISO) criteria for accuracy.